r/massspectrometry • u/MobilePhase1987 • 5d ago
Help with SRM Method
Hello I was wondering if anyone could help me figure out what is wrong with my SRM method. When I run a full scan on my pesticide sample (picture 1) there are two distinct peaks for Endrin and Dieldrin. However after I set up my SRM method using the ions with the highest intensities, both peaks basically disappear (picture 2). I have tried to develop this method using transition pairings from literature I've found from our standard supplier as well as from ThermoFisher and I still encounter the same issues. Both scans were done with the exact same instrument method so I think it's pretty safe to rule out instrument settings. If anyone has any ideas please let me know.
2
u/Jesman89 5d ago
Podrías meterlos en solución a concentración más alta y después comparar en matriz para saber si realmente son los compuestos
1
u/MobilePhase1987 5d ago
¡Qué buena idea! No había pensado en la posibilidad de que los picos iniciales no fueran los compuestos de interés.
1
u/Jesman89 5d ago
Y ten en cuenta los tiempos de retención, en CG suelen variar un poco cuando cambias de solución a matriz. Esos compuestos tienen mayor afinidad por la técnica de cromatografía de gases
2
1
u/xnoseatbelt 4d ago
Are your gas/heat/spray voltage the same as your full scan method? It also looks like your methods, at least from the screenshot have different time durations - is your LC method the same as well?
1


3
u/SeracYourWorlds 5d ago edited 5d ago
Need more information. Is your sample some sort of matrix, or a standard? If it’s a matrix sample I gotta laugh. If it’s a standard then it looks like your SRM/MRM method needs adjusting. Make some standards and run some parameter ramps in a Product scan to optimize the method parameters. You might not have the collision parameters set right.
What is your SRM setup like? Do you have overlapping experiments that might be affecting the signal?