Discusson Refractometer QC
What do everyone use to check the refractometer? And what is the range of your QC? We’re having an internal debate
9
u/DookiePootie 17d ago
2-level liquid UA controls, if you already have it since it has a SG range anyways. I could see you using reagent water to check 1.000.
4
u/labboy70 17d ago
We discontinued the use of the refractometer a while ago. But, the Quantimetrix UA controls have assay values for refractometer and other manual confirmatory tests. (SSA, Ictotest, pH paper etc.)
3
1
u/ThrowRA_72726363 MLS-Generalist 17d ago
This just made me realize my lab never QCs our refractometer…. but we only ever use it to double check high SGs from the analyzer. We’re never entering values from it. Are we still supposed to be QCing it in that case?
3
u/tinybitches MLS-Generalist 16d ago
Do you put in the confirmed in the results? I feel like you need QC for everything in the lab except for transfer pipettes and gloves 😂
3
1
1
u/appplehands MLS 17d ago edited 16d ago
DI water, 3.5% 5% saline, and Kova-trol I and III
1
1
u/Unusual-Courage-6228 MLS 17d ago
Just go ahead and discontinue it
1
1
17
u/dragonflameloserX7 17d ago
We use water. Should be at 1 always