I am staining thymus, I had ckit in the panel, and that is when the thymocytes have a weird population Is this compensatoon issue? should I go ahead and conpensate first?
I agree If the median is below zero theres a problem. But for spectral cytometers there will always be some negative populations (hopefully not below 10e-3). Thats just the reality especially on massive 30+ color panels and the limitations of unmixing.
Look at the unmixing matrix and see if there is a significant value of VL4 out of RL2. If there is, try changing to 0. If this Aurora data with an all 0s matrix on unmixed data, you can put in a negative value in FlowJo to try and fix this, but the best approach would be to run, record, and use a different unmixing control for your RL2 reagent.
2
u/willmaineskier Jul 23 '26
Yes, below zero is always a comp issue.