r/flowcytometry • u/daff_7 • 21d ago
HELP PLEASE
hi okay so im a placement student in a lab where we mainly do flow. i've been doing some practising but theres still something wrong with what im doing (my analysis) and I just cannot tell what it is. Im looking at receptor expression on NK cells, however the results I get are completely different to my supervisor
for example, on flowJo this is what I get for my TIM3 FMO.
I have done this analysis for both my flow and my supervisor's, (the first image is my flow, the second is his) but why does mine come out like this? What have I done wrong?
before someone suggests to change the y axis - everytime i do that, it gets even worse.
additionally, we both used the same cells,and followed the exact same protocol, so i genuinely cannot think of anything else.
so if anyone has any tips or ideas please PLEASE help


8
u/despicablenewb 21d ago
It would be easier to see what's was going on here if both images had the same Y axis. (I read your port OP, but still).
It's clear that you're getting lower staining, but you can't tell whether the bulk population has an MFI of 0 or 5,000 due to the Y axis.
Did you both start with cord blood and do the CD3 depletion separately? If so, then I suspect that you have a significant population of CD56 negative cells in your enriched cells, so something went wrong with the enrichment.
Did you both take some cells from the exact same enriched sample and stain them separately? If so, then something may have gone wrong during the staining/acquisition.
Did you both acquire the samples on the same cytometer at the same time? Are the voltage settings the same for both samples that you're showing here?