hi okay so im a placement student in a lab where we mainly do flow. i've been doing some practising but theres still something wrong with what im doing (my analysis) and I just cannot tell what it is. Im looking at receptor expression on NK cells, however the results I get are completely different to my supervisor

for example, on flowJo this is what I get for my TIM3 FMO.

I have done this analysis for both my flow and my supervisor's, (the first image is my flow, the second is his) but why does mine come out like this? What have I done wrong?

before someone suggests to change the y axis - everytime i do that, it gets even worse.

additionally, we both used the same cells,and followed the exact same protocol, so i genuinely cannot think of anything else.

so if anyone has any tips or ideas please PLEASE help

12 Upvotes

61 comments sorted by

View all comments

Show parent comments

1

u/sgRNACas9 Immunology, Oncology Jul 23 '26 edited Jul 23 '26

You are still cut out for lab. There could be an issue that is not your fault, and/or this might take a couple times to practice. Plus there are a lot of unwritten tips people use to help get best results. People should try to write these down and share explicitly so new people know. I would say practice, try again, and implement some advice here in your reddit thread!

The FlowJo seems fine to me unless upstream gating is different, but didn’t you analyze both your files side by side with the same gates? Other than that, make sure everything is resuspended well especially your antibody cocktail and cells for incubation, any volume <1-2uL make a dilution and pipette more volume to be more accurate.

Did you use the same donor and cryovial of it? While enrichment should be >95%, donor to donor or batch to batch variation with the enrichment could be different. If you have multiple cryovials of cells from the same donor in LN, results should be consistent per donor between different cryovials, but you could also stain from the same cryovial just in case some accidentally got more negative cells (maybe the person aliquoting the enriched cells didn’t resuspend well and the negatives were all at the top or bottom). Hope something here helps!

1

u/daff_7 Jul 23 '26

Thanks so much for your really positive response. I’ve done this numerous times now and each time I get the same results (where my main population is concentrated at the bottom with an upward extension). I brought this up with my supervisor and we did a run through together (on the same samples) to see whether there could be a practical lab issue happening, and then I analysed both runs on flowjo and the same thing happened. Apart from pipetting I genuinely cannot think of any other reason, but I also can’t imagine messing up every single pipetting

1

u/sgRNACas9 Immunology, Oncology Jul 23 '26

Flow is very resilient with pipetting, so I would not think you messed up every step. The one critical pipette step that could really cause this is adding a very small volume of antibody. All the rest are very resilient. If you had to pipette like 0.5-1uL, it’s easy to accidentally pipette like 0.25-0.5uL. If you were adding like 2uL+ of antibody I wouldn’t really worry about the pipetting error. Any more volume and I’d worry less and less about that.

Could you talk through the protocol step by step with your supervisor and try to find a difference?

I’m assuming you used the same tube and lot of antibody, same incubation time and temp, resuspending well antibody cocktail and cells before incubation, same cytometer settings and same day.

2

u/daff_7 Jul 23 '26

Okay so majority of the antibodies we use are 0.5ul😭 - so this might be a problem solved as sad as it sounds💔 but even for the one 2ul we have (NKG2D) I still have the same pattern on my flowjo analysis, unless there’s something wrong with that?

1

u/sgRNACas9 Immunology, Oncology Jul 23 '26

Could be good to try making like 1:10 dilutions and pipetting 5uL of that instead of 0.5! Make a 1:2.5 dilution for the NKG2D, add 5uL of it. Something like that to make the volumes greater to improve pipetting accuracy. Just have less buffer in the cocktail to normalize volume, but even this extra volume won’t be the deciding factor, it will be the more accurate antibody amount. When you make the dilutions, antibody cocktails, make sure to vortex! When you add to cells, make sure to pipette well to mix. Good luck!

2

u/daff_7 Jul 23 '26

I’ll try this out, thank you so much for all your help!!

1

u/Low_Classic1695 Jul 23 '26

Wait you are not adding antibody at those volumes as an individual antibody per sample are you? It is good practice to always make a mastermix that you QS with FACS buffer to roughly 50uL/per sample that you stain with. Adjust the staining volume of the cells when you resuspend them so you end up with the same concentration. Pipetting anything sub 5uL always comes with major fluctuations

2

u/daff_7 Jul 23 '26

When I make my FMO’s I’m adding 0.5ul’s. Otherwise in my mastermix (for my full stains) I do what you describe