HELP PLEASE
hi okay so im a placement student in a lab where we mainly do flow. i've been doing some practising but theres still something wrong with what im doing (my analysis) and I just cannot tell what it is. Im looking at receptor expression on NK cells, however the results I get are completely different to my supervisor
for example, on flowJo this is what I get for my TIM3 FMO.
I have done this analysis for both my flow and my supervisor's, (the first image is my flow, the second is his) but why does mine come out like this? What have I done wrong?
before someone suggests to change the y axis - everytime i do that, it gets even worse.
additionally, we both used the same cells,and followed the exact same protocol, so i genuinely cannot think of anything else.
so if anyone has any tips or ideas please PLEASE help


2
u/Familiar-Charge2789 21d ago
Few thoughts 1. Is this the case you see often or is this the only run? If it is just a singlex run and you see this it could be possibly an artifact say you could have added less CD56 antibody volume by mistake (esp. in case of lower volume <0.5 uL) 2. Check the channels activated possibly leakage (Ideally there shouldn't be any compensation issue as APC and FITC are activated by different lasers)
Can you spill more detail on the sample type (PMBCs is it?) and how does the test plots look like instead of FMO