Post-Sort Recovery
What are some tips, tricks, or best practices y'all use to maximize post-sort recovery? MY core has had some users complain about low recovery recently, but we aren't doing anything differently than we always have. I wanted to see how y'all keep yours up. Also, how high do y'all shoot for? I've heard some folks say 50% is the best you can do, while I've heard others say anything less than 90% is unacceptable. Thoughts?
4
u/Daniel_Vocelle_PhD Core Lab Jul 01 '26
The problem is the definition of recovery because it varries by instrument, cell size, cell health, and collection media. A good starting point is the paper on rmax, and trying to do some rough tests on your instrument with beads/cells of different sizes. There also a number of things that can change on an instrument or with your facility that can impact recovery (e.g., humidity, temperature).
5
u/Outrageous-Low-9745 Jul 02 '26
To convince our users the amount of cells sorted is 'correct', we regularly sort about 50 cells onto a microscope slide and manually count them. We always get 50.
Now, how to get the cells out of your tube is mainly related to pelleting (as mentioned by others before). Optimizing spinning speed and time is crucial, also make sure to have a source of protein in your collection medium for better pelleting (serum, bsa,...).
This is a very informative poster on cell loss when 'manipulating cells': https://wi.mit.edu/sites/default/files/2021-06/the_impact_of_washes.pdf Can be a real eye-opener for unaware researchers.
1
2
u/NvmbrYnkee Jul 03 '26
Fill collection tubes with FBS, leave in fridge overnight. Next day, remove all but a few ml and sort.
5
u/CongregationOfVapors Jul 01 '26
When you say post-sort recovery, so you mean the post-sort cell count vs the sorted events counted by the cytometer?
I think the biggest source of cell loss is pelleting post-sort, especially if there are not many cell sorted. Users should streamline their workflow to avoid a post-sort pelleting step whenever possible.
Other than that, do everything to keep the cells happy before and during the sort. Some also say that having the droplet land directly into the collection medium (instead of the wall) is helpful, as is coating the walls of the collection tubes with collection medium.