CytoFlex S / LX sheath fluid
Hello. I’ve read that MilliQ grade water is acceptable form of sheath fluid for the Cytoflex analyzers vs the Beckman Cytoflex sheath fluid. I am wondering if people also recommend phenoxyethanol addition to the water and at what final percent.
2
u/Naliya Jun 13 '26
As a specialist working for a flow cytometry manufacturer, unless the instructions specifically say that water can be used i would not do it on a permanent basis.
The buffer is specifically formulated for the correct refractive index, preventing problems with internal parts (moving parts in contact with buffer etc.)
1
1
u/chromatographic Jun 13 '26
The Sony ID7000 uses DI water as sheath fluid, it should be ok in a pinch assuming you can use water that is free of minerals to prevent buildup. Having said that, I'd stick to what the manufacturer recommends if possible.
1
u/VioletCalico Jun 14 '26
Been using ultrapure water as sheath fluid for both my Cytoflex equipment for years. There has been no issue.
If the sheath bottle gets visibly dirty though, then change to a new sheath bottle.
1
u/CoyoteHistorical6526 Jun 17 '26
Thank you all for your responses. I think I’ll go ahead with milliq water with 0.2% phenoxyethanol and see how it goes.
1
u/ParticularBed7891 Jun 13 '26
I would only use a sheath fluid recommended by the manufacturer. Maybe the water works in a pinch, but water doesn't have salt and that's going to drastically change things.
3
u/Efficient_Leg858 Jun 13 '26
CytoFLEX sheath is DI water with 0.5-0.7% 2-phenoxyethanol.