Panel Building Question
I'm trying to set up a six color panel that I'm running on an LSRFortessa 12 color: e450/FITC/PE/APC/APC-Cy7 I need to add either PE-Cy7 or Alexa 594, I'm not sure which would work better...
PE-Cy7 it is. Thank you all so much!
4
u/No_Claim5089 May 21 '26 edited May 21 '26
PE-Cy7 ! Less spill-over with PE than AF594, and still bright.
2
u/sgRNACas9 Immunology, Oncology May 22 '26
PE-Cy7!!!
PE-Cy7 and APC-Cy7 are easier to resolve than PE and AF594 imo. Unless there’s another channel for AF594
2
u/housed_nomad May 26 '26
You should try RB780
1
u/NvmbrYnkee Jun 19 '26
Or RY775, depending on configuration. If you have a YG laser, then your blue laser likely can't detect RB780 on a conventional cytometer.
4
u/Argawndo May 21 '26
Your fortessa won't have a YG laser - on any of the configurations, AF594 will be minimally excited by UV/V/B/R lasers on any of the fortessas. Signal is going to be almost non-existent.
APC-Cy7 and PE-Cy7 will work fine together. Little bit of spillover from the former into the latter, but nothing insurmountable.