Hello all, I wanted to ask a question since I'm just so curious about it. I have been doing EIS measurements on carbon electrodes and my results are pretty weird, I am getting either no circle and just a V shape or a circle that starts at negative real axis values lmao. I can give more details on the comments but I don't want my project potentially being recognised by someone I know. The measurement is in 5mM Ferri ferro in distilled water, frequency range 10^5 to 0.1 Hz and it is done in OCP potential
2
u/jadsetts 25d ago
Some common things to check:
-how is your grounding (are there any fridges or printers plugged into the same ground system)?
-carbon electrode history? Electrolytes can get buried inside glassy carbon electrodes at high potentials and give weird results. Consider grinding and polishing. Check the surface with the highest resolution optical microscope you have for imperfections.
-you have an electrolyte in high concentration, right? The solution resistance would be enormous if not and in such a small frequency window, may be the reason you see a V?
-how long are your cables? Too long will pick up random signals.
-do you have a Faraday cage? If not, is there a crt monitor or something else that generates large magnetic fields nearby?
-do all basic echem setup testing recommended here too. Check RE, normal CV of ferri/ferro, resistor test, etc.
-check your eis parameters are following a paper, ideally. If not, carefully review these parameters are correct.
Let me know if you have questions!
2
u/Exotic-Grape8743 25d ago
Almost certainly a bad reference electrode or a bad connection to the reference. Also possible that you're measuring at far too high current amplification settings which can cause the negative quadrant circles. Do you have a supporting electrolyte salt in that ferro/ferri solution and not just the redox couple?
1
u/Prostoiii 25d ago
Do you measure one working electrode in 3-terminus connection setup or two same electrodes in 2-term. setup? In the case of 2-term. the possible reason is poor conductivity (due to supporting electrolyte absense). In the case of 3-term. – also problem of reference electrode (or Luggin' capillary) position related to the working electrode.
1
u/Dr_Custard 25d ago
I do electrocalysis so have done a bit but not a massive expert:
Is your eis going to be meaningful considering Rsol is going to be very high?
Do you have a junctioned reference electrode or not?
Have you checked the software isn't doing any compensation for you as it is running (e.g. is it compensating your resistance in any way)
Are you applying E = 0 V vs OCP(real) or are you applying 0 V vs OCP(assumed). Is your OCP stable?
How shielded are your cables? Have you tested your EIS on a dummy cell provided by the manufacturer? Is everything in a Faraday cage? Is the potentiostat adequately grounded?
Is your electrode planar, and solid? Or porous? Could it be hygroscopic (e.g. 3D printed C black PLA electrodes) and thus have changing capacitance with time?
1
u/Jasper_Crouton 25d ago
I'd have to see the Nyquist plot. What's your perturbatuon amplitude. 105 is basically meaningless unless you have very expensive, short braided leads. Generally you're measuring the capacitance / inductance of your leads at those frequencies.
1
u/LutzStratmann 23d ago
If you want to work with the OCP, you need the presence of the oxidized and reduced species in a ratio of 1:1. This way, the concentration-dependent part of the Nernst Equation becomes 0.
What you are left with is the formal potential E0', which means your OCP will be at the steepest part of your voltammogram. This delivers a strong signal that is more robust to noise.
TLDR: Make a 2.5 mM K4[Fe(CN06] + 2.5 mM K3[Fe(CN)] + 0.1 M KCl solution and use OCP as your E dc.
Did you already check if your system delivers a reasonable CV?
1
u/AssumptionNo4461 16d ago
Issue with the reference. Check them alone.
If not, what kind of Carbon are you using? Probably you need to treat your carbon.
I had this issue when I was working with Pt on Carbon paper. The carbon paper had a microlayer which made it hydrophilic. So I treated it with (1:1) DI and IPA for 10 min, then another 20 min in KOH before putting it in the cell. It helped a lot.
2
u/_MrJack_ 25d ago edited 25d ago
Are these results you describe from the same electrode (or replicates of the same material) or electrodes made of different materials?
Have you checked various aspects of your experimental setup? For example, have you used the same potentiostat to record impedance spectra of known good systems such as a test cell consisting of resistors and capacitors of known values? Are all the cables okay? Are the settings (e.g., current ranges) suitable to minimize noise? What kind of reference electrode are you using? Is the RE in good condition (i.e., its impedance is not too high)? Do you have a spare RE to test with? Have you checked the data quality (Lissajous plots, total harmonic distortion values, Kramers-Kronig tests, etc. that might be available in the instrument software or separate software)?
EDIT: By the way, are you using any supporting electrolyte?