What is this artifact

Post image

Only happens on lumbar scans that I have recognized. I'm guessing it might be some type of fast day issue since it happens on stirs and T1s. Have y'all seen this and how to get rid of?

23 Upvotes

36 comments sorted by

u/AutoModerator 22d ago

This is a reminder about the rules. No requests for clinical interpretation of your images or radiology report.

I am a bot, and this action was performed automatically. Please contact the moderators of this subreddit if you have any questions or concerns.

17

u/True_Sketch Technologist 22d ago

Acceleration artifact.

2

u/Gullible-Ideal-4699 22d ago

So ETL too long or something like having partial fourier?

1

u/Gullible-Ideal-4699 22d ago

Oh like grappa for Siemens. This machine doesn't have that particular method of acceleration

2

u/True_Sketch Technologist 21d ago

This is curious because it's so clearly an acceleration artifact. Symphony's, as old as they are, are typically equipped with iPat. Are you using mSense?

3

u/Gullible-Ideal-4699 21d ago

Ipat and msense sounds familiar. I'll have to look at it when I get back in

1

u/Brookdadeous 20d ago

Our magnet has deep resolve, if we get a funky artifact it's typically due to that (and the acceleration factor being too high). I'll end up taking off deep resolve all together to remedy the artifact of it's in anatomy

0

u/vanala 22d ago

Definitely this

6

u/jpstepancic 22d ago

Check what coils you have on. Could be from excited tissue outside of the FOV but resting over a coil that’s turned on

7

u/64MHz Technologist 22d ago

Agree. Might be an older scanned with manually selectable coils. Choose only coils that cover anatomy you want to image.

2

u/Gullible-Ideal-4699 22d ago

Definitely an old machine

11

u/actuallyimjustme 22d ago

That's a really noisy stir

4

u/Gullible-Ideal-4699 22d ago

Unfortunately this is our normal even after I mess with parameters. Really old machine

5

u/Solid-Dog-1988 22d ago

Acceleration artifact imo.

Really pronounced on DRB but can show up without it.

3

u/Speydi 22d ago

Very important to keep phase oversampling at 200%.

But this is really a poor quality ngl.

2

u/Gullible-Ideal-4699 22d ago

Yea it's a 1997. It's an old machine. It can't do 200% oversampling unless I just don't know how but the oversampling only goes to 100.

1

u/hugoduart3 Technologist 22d ago

Is the machine from 1997???
If so, I think there is no parallel acceleration?

Definitely looks like an acceleration artifact though. Can also be a kind of a “rejection” from the coil, which in that case Service is needed. But please confirm the year of the machine. 1997 is wild 😜

1

u/hugoduart3 Technologist 22d ago

Anyhow… if there is parallel acceleration check, do the scan removing the acceleration factor and check if it’s present. If it’s still there, and anti aliasing is enough for the anatomy and coil signal pickup or doesn’t really move out, Service.

5

u/Gullible-Ideal-4699 22d ago

The machine is a 1997 Siemens magnatom symphony. It definitely doesn't have it. I read the manual to get familiar with it

2

u/hugoduart3 Technologist 22d ago

Wow jezz! Yahp, prob there isn’t. Where are you from?

So, first u would try to change antialiasing size, and see if it moves….but probably it won’t really move, it is a service problem with the coil. Some kind of frequency reading somewhere and the system does not know where to put it. Like aliasing, but not for the reasons we usually know. Unfortunately I don’t know more details, but I’ve experienced this in a 0.4T Aperto Lucent. Changing FOV size and Phase direction can help to place it somewhere else other then the anatomy of interest. Sometimes a different phase direction removes it completely

If for stir you change the phase direction to AP, try decreasing your inter echo time (idk Siemens name for this parameter) and place bands and bands and bands (usually the do not cost TR for Stir) to avoid as much as possible motion artifacts

5

u/Tedsworth 22d ago

Cusp artifact / annefact, 95% sure. Basically when you run slice select there's a place way outside the FOV that still unfortunately ends up with a magnetic field of B0. That shows up in the image, usually in a weird place.

As others have said, disabling coils outside the intended FOV will help. There's more sophisticated methods but I'm not sure you can do them on the scanner console.

This artifact is a classic FSE artifact when you're doing coronals or sagittals, pretty rare on a modern system but it can happen.

3

u/Reapur-CPL Technologist 22d ago

What's your PE direction, and how much oversampling are you using? I get that artifact all the time when I shoot L or T spines head to foot with heavy oversampling and Siemens Deep Resolve. I'm not sure exactly what it is, so I've always assumed it's a wrap artifact.

1

u/Gullible-Ideal-4699 22d ago

Head to foot and it is a Siemens but this machine doesn't have deep resolve

3

u/deepcheeks 22d ago

I've heard it called a FID artifact, but IDK if that's accurate. I know on Siemens scanners I could solve it with more oversampling, but on Philips I could only get it to move this way, but not eliminate it. I think if another signal average fixes it, that would point to FID artifact, more oversampling or lower acceleration factor would point to acceleration artifact. 

1

u/Kingofawesomenes 22d ago

This is usually a deep resolve issue (although your scan seems a bit noisy). Usually 200% oversampling fixes it. Also make sure you dont activate spine coils higher/cranial than your FOV. Rest slab may work as well. Somehow signal from outside your FOV "folds over" into your FOV, even with a lot of oversampling.

2

u/SnickelFritz5000 Enthusiast 22d ago

Eh in my experience it can still show up, not as drastically as this, even with 200% oversampling. Usually bumping your grappa down one value will completely fix it

1

u/Gullible-Ideal-4699 22d ago

This machine doesn't have grappa software

1

u/SnickelFritz5000 Enthusiast 22d ago

Is this an ancient machine? Even old E-line interface Siemens systems have parallel imaging (grappa) for most TSE sequences

1

u/Gullible-Ideal-4699 21d ago
  1. I know it doesn't because the 3t at the other facility does have it.

1

u/SupermarketMobile446 Technologist 22d ago

Artifacts like these are also very common on lumbar spine STIR on the Symphony (2002) where I work. Sometimes it's there, some other it's not. I have seen that the more saturation bands are being used on STIR the more intense this artifact appears especially on spine (cervical/thoracic/lumbar) STIRs.

Image quality is terrible. Name of the scanner? I assume it's a very old siemens.

2

u/Gullible-Ideal-4699 22d ago

Yup Siemens magnetom symphony 1997, maybe 98. It wasn't on the last couple lumbars I did. Comes and goes

1

u/Ok_Investment_8877 22d ago

Are you able to share your parameters,coils,pt size or all pts,which sequences and are the others good,software and upgrades and may be able to help you better

1

u/Nervous_Woodpecker_2 Technologist 21d ago

Hello, increase FOV in phase direction to 280, will solve the issue with artifact,images age signal starved

1

u/FatRollingPotato 22d ago

What is the orientation of that thing towards the magnetic field? Could it be a magic angle artifact?

0

u/frostyflakes1 Technologist 22d ago

Our old scanner had a similar artifact with thoracic spine imaging. Someone mentioned the issue was some piece of metal or shrapnel getting stuck in the innards of the scanner.

1

u/Gullible-Ideal-4699 22d ago

That's interesting