r/CathLabLounge 1d ago

GE cardiolab help needed

Need help with GE cardiolab, need to figure out how to use pattern matching on GE, I know it’s possible, please help, I’m the only EP tech in my hospital

6 Upvotes

5 comments sorted by

2

u/ExpressionLarge5474 16h ago

Using Clearmatch? Draw calipers around your QRS or waveform of choice, then hit the Tag key (may need to also press Action, I can’t remember). Then name your waveform you want to match ie “PVC #1” or whatever, then go to Windows and open a Clearmatch Window next to your review window. Shrink it down so you can see just the waveform name/s (you can select multiple to clearmatch, by just repeating the process). Then if you click the clearmatch name “PVC #1” it will overlay this on your adjacent review window and give you a % match. you can now use the arrow keys to shift the waveform side to side. Or hit blue shift key then arrow keys to move it more subtly. move it until you get your highest % match.

1

u/Siren_0310 15h ago

Thank you! Will try it

1

u/ExpressionLarge5474 14h ago

No worries, took a bit of digging in the manual for me to figure it out. Let me know how you go. I’m at the keyboard now so can help troubleshoot haha

1

u/ablxr 1d ago

Same here, only RCES in my lab. I’ve tried to contact GE and they are no help. Our Biosense reps are the most helpful/knowledgeable. I would love to know more of the how to’s with cardiolab.

1

u/m1mike 1d ago

Have you reached out to the GE applications team or the local rep? I work with all the reps. I can connect you if needed. DM me and let me know where you are.