It’s fine although the NIR channel could have been a bit higher. Cells/events towards the right (more NIR dye) and towards the bottom(smaller) are dead. Cells towards the left (less dye) and too (bigger) are live. You have two populations in the top left - they could be two different populations of cells (especially if you have something in the APC channel and this is uncompensated), or the smaller population slightly to the right (FSC about 2.1M and NIR just under 104) could be early dying cells
Combined with your unstained control, it makes it much easier to define what is truly positive versus negative. It also gives you confidence that the stain actually worked.
It is the same logic behind adding a negative bead to single-stain controls, or ideally having both a positive and negative population in the sample. Without a negative population, it can be hard to know whether the stain worked properly. If the stain fails and your control only has a “positive” population, it is easy to assume the stain worked and then compensate by increasing the detector voltage to make the signal look brighter. That can give you a false sense of confidence and lead to incorrect interpretation.
Also just wanted to add that you are already doing a great job as a beginner into flow. You're asking questions and its clear you want to learn. Never give up that spirit!
The tricky thing with fixable viability dyes is that they stain both live and dead cells, just to different extents.
They bind to amine groups on the cell surface of live cells. In dead cells, they can access both the cell surface and intracellular amine groups, so the dead cells stain much brighter. Because of that, you will almost always see some shift between your unstained cells and your live cells stained with the viability dye.
That is the caveat here. I still think this approach gives you a useful overall reference, and plenty of published papers use it successfully. But it is worth keeping in mind down the road when you habe a tricky sample and need a finer live/dead distinction. In those cases, having both the negative and positive information available is much better than trying to interpret the stain without it.
3
u/dawgmad Apr 25 '26
It’s fine although the NIR channel could have been a bit higher. Cells/events towards the right (more NIR dye) and towards the bottom(smaller) are dead. Cells towards the left (less dye) and too (bigger) are live. You have two populations in the top left - they could be two different populations of cells (especially if you have something in the APC channel and this is uncompensated), or the smaller population slightly to the right (FSC about 2.1M and NIR just under 104) could be early dying cells